Sterile Processing Technician Exam
Decontamination Practice Questions
39 practice questions with detailed explanations — aligned to the Sterile Processing Technician Exam.
Master Decontamination to boost your score on the Sterile Processing Technician Exam. Each question below mirrors the style and difficulty of real exam questions, complete with detailed explanations so you understand the why behind every answer. Work through all 39 questions, review any that trip you up, and use the related topics below to round out your preparation.
Q1.What is the first step in the decontamination process?
A.Manual cleaningB.Point-of-use treatmentC.SortingD.Disinfection✓B. Point-of-use treatmentExplanation: Decontamination begins at the point-of-use (in the OR) to prevent blood/soil from drying on instruments. This makes subsequent cleaning steps effective.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q2.The ideal temperature for enzymatic cleaners is typically:
A.Above 140°F (60°C)B.Below 100°F (38°C)C.Between 109°F - 140°F (43°C - 60°C)D.Boiling✓C. Between 109°F - 140°F (43°C - 60°C)Explanation: Enzymes are proteins that work best at specific warm temperatures. Too hot (>140°F) kills the enzymes; too cold renders them sluggish. Always check manufacturer IFUs.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q3.Cavitation is the cleaning mechanism used in:
A.Washer-disinfectorsB.Ultrasonic cleanersC.Cart washersD.Steam sterilizers✓B. Ultrasonic cleanersExplanation: Ultrasonic cleaners use sound waves to create microscopic bubbles that implode (cavitation), pulling soil off surfaces and crevices.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q4.Which type of water is recommended for the final rinse of critical medical devices?
A.Tap waterB.Critical water (Deionized/Distilled/Reverse Osmosis)C.Softened waterD.Utility water✓B. Critical water (Deionized/Distilled/Reverse Osmosis)Explanation: Final rinses require water free of minerals and impurities (Critical Water) to prevent staining, corrosion, and pyrogen contamination.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q5.When manually cleaning a device, the brushing motion should be performed:
A.Above the water line to see clearlyB.Under the water line to prevent aerosolizationC.Vigorously with a wire brushD.Only on the exterior surfaces✓B. Under the water line to prevent aerosolizationExplanation: Brushing under the water surface prevents aerosols containing pathogens from becoming airborne and contaminating the technician.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q6.A washer-disinfector uses which process to achieve thermal disinfection?
A.Chemical exposureB.ImpingementC.Hot water rinse (typically 180°F - 203°F)D.Ultrasonic waves✓C. Hot water rinse (typically 180°F - 203°F)Explanation: Thermal disinfection in automated washers relies on sustaining a high temperature rinse for a specific time to kill vegetative microorganisms.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q7.Items with lumens (channels) must be:
A.Soaked onlyB.Brushed with a properly sized brush and flushedC.Processed in the ultrasonic onlyD.Sterilized without cleaning✓B. Brushed with a properly sized brush and flushedExplanation: Soaking isn't enough. Friction (brushing) is required to remove soil inside the lumen, followed by flushing to clear debris.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q8.The pH of a detergent used for anodized aluminum containers (rigid containers) should be:
A.High alkalineB.Neutral (pH 7)C.AcidicD.Low alkaline✓B. Neutral (pH 7)Explanation: Anodized aluminum is damaged by high/low pH. Neutral pH detergents are required to prevent corrosion/pitting of the protective layer.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q9.Proper airflow in the decontamination area should be:
A.Positive pressureB.Negative pressureC.Neutral pressureD.It doesn't matter✓B. Negative pressureExplanation: Negative pressure ensures that air flows *into* the decon area from clean areas and is exhausted outside, preventing contaminants from drifting into clean prep areas.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q10.TASS (Toxic Anterior Segment Syndrome) is an acute inflammation of the eye caused by:
A.BacteriaB.Inadequate removal of detergent or chemical residues from ophthalmic instrumentsC.Viral infectionD.Patient allergies✓B. Inadequate removal of detergent or chemical residues from ophthalmic instrumentsExplanation: TASS is a sterile inflammatory reaction caused by toxic residues (like detergents, endotoxins, or enzymes) left on eye instruments. Thorough rinsing with critical water is essential.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q11.Before placing instruments into an ultrasonic cleaner, they must be:
A.SterilizedB.Grossly cleaned of visible soilC.DryD.Lubricated✓B. Grossly cleaned of visible soilExplanation: Ultrasonic cleaners are for fine cleaning. Gross soil inhibits cavitation and contaminates the bath quickly. Pre-cleaning is mandatory.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q12.Descaling is the process of:
A.Removing instruments from a trayB.Removing scale (mineral deposits) from the chamber of a washer or sterilizerC.Checking weightD.Lubricating instruments✓B. Removing scale (mineral deposits) from the chamber of a washer or sterilizerExplanation: Hard water leaves chalky mineral deposits (scale) on equipment walls. Descaling agents (acidic) remove this buildup to maintain efficiency.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q13.Which agency regulates disinfectants and sterilants used on non-living surfaces?
A.FDAB.EPA (Environmental Protection Agency)C.OSHAD.CDC✓B. EPA (Environmental Protection Agency)Explanation: The EPA regulates surface disinfectants and liquid chemical sterilants (as pesticides). The FDA regulates high-level disinfectants used on medical devices.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q14.Personal Protective Equipment (PPE) for the decontamination area includes:
A.Lab coat onlyB.Gloves, fluid-resistant gown, eye protection, face mask/shield, and shoe coversC.Sterile gloves and gownD.Street clothes✓B. Gloves, fluid-resistant gown, eye protection, face mask/shield, and shoe coversExplanation: Full PPE is required to protect against splashes and biohazards. Heavy-duty utility gloves are preferred over exam gloves.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q15.A 'High-Level Disinfectant' (HLD) kills:
A.All microorganisms including high numbers of bacterial sporesB.All vegetative bacteria, viruses, fungi, and mycobacteria, but not necessarily high numbers of sporesC.Only bacteriaD.Only viruses✓B. All vegetative bacteria, viruses, fungi, and mycobacteria, but not necessarily high numbers of sporesExplanation: HLD kills almost everything (TB, viruses, fungi) but cannot guarantee the kill of high numbers of spores (that requires sterilization).
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q16.The minimum effective concentration (MEC) of a high-level disinfectant must be tested:
A.Once a weekB.Once a dayC.Before each use (or load)D.When the bottle is opened✓C. Before each use (or load)Explanation: HLD solutions (like Glutaraldehyde or OPA) degrade over time. You must dip a test strip before *every* use to ensure potency.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q17.According to the Spaulding Classification, 'Semi-critical' items require at least:
A.Cleaning onlyB.Low-level disinfectionC.High-level disinfectionD.Sterilization✓C. High-level disinfectionExplanation: Semi-critical items touch mucous membranes (e.g., endoscopes, anesthesia blades) and require High-Level Disinfection. Critical items (enter sterile tissue) require sterilization.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q18.Why must instruments be disassembled before cleaning?
A.To count the partsB.To ensure cleaning agents contact all surfacesC.To make them fit in the trayD.It is not required✓B. To ensure cleaning agents contact all surfacesExplanation: Bioburden hides under movable parts. Disassembly ensures detergent and water reach every surface.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q19.Ortho-phthalaldehyde (OPA) is a common:
A.Enzymatic cleanerB.High-level disinfectantC.Sterilant gasD.Lubricant✓B. High-level disinfectantExplanation: OPA is a widely used liquid high-level disinfectant for flexible endoscopes. It is a safer alternative to glutaraldehyde.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q20.If a rigid container is returned to decontamination with the filter still in place:
A.The filter can be reused if cleanB.The filter must be removed and discarded, and the retention plate cleanedC.The container can be washed with the filterD.Leave it for the assembly side✓B. The filter must be removed and discarded, and the retention plate cleanedExplanation: Disposable filters are single-use. They must be removed in decon to allow proper cleaning of the lid and retention plate.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q21.When loading a mechanical washer, instruments should be placed:
A.Stacked on top of each otherB.In the closed positionC.In the open/unlocked positionD.In peel pouches✓C. In the open/unlocked positionExplanation: Box locks and ratchets must be open to allow spray action to reach the hinge areas and jaws.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q22.The 'hollow' cavitation test used for ultrasonic cleaners involves:
A.A ceramic discB.Aluminum foilC.A TOSI testD.A biological indicator✓B. Aluminum foilExplanation: The aluminum foil test detects uneven cavitation. (Note: Newer commercial tests like SonoCheck are also used, but foil is the classic textbook answer for verifying cavitation energy distribution).
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q23.Prions (which cause CJD) are uniquely resistant to:
A.Standard cleaningB.Standard sterilization cycles (normal steam/EtO)C.High-level disinfectionD.All of the above✓D. All of the aboveExplanation: Prions are misfolded proteins that are incredibly resistant. They require extended sterilization cycles (e.g., 18 min at 273°F or 1 hour at 250°F) and specific cleaning protocols.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q24.Which detergent is best for removing stubborn organic soils like dried blood?
A.Acidic detergentB.Enzymatic detergent (Protease)C.High alkaline detergentD.Neutral detergent✓B. Enzymatic detergent (Protease)Explanation: Protease enzymes specifically break down proteins (blood/tissue). Lipase breaks down fats.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q25.Transporting contaminated items to the decontamination area requires:
A.Carrying them in handB.A closed, leak-proof container marked with a biohazard labelC.An open wire cartD.A clear plastic bag✓B. A closed, leak-proof container marked with a biohazard labelExplanation: OSHA requires biohazardous materials to be contained in leak-proof, labeled containers to protect staff and the environment during transport.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q26.Power equipment (drills, saws) should generally be:
A.Immersed in waterB.Cleaned manually (wiped down) and never immersedC.Processed in the ultrasonicD.Put in the washer-disinfector✓B. Cleaned manually (wiped down) and never immersedExplanation: Immersion allows water to enter the motor housing, destroying the mechanism. Follow IFU carefully—usually manual wipe-down.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q27.Biofilms are:
A.A type of soapB.A protective slime layer created by colonies of bacteria that resists cleaning and disinfectionC.A sterile barrierD.A lubricant✓B. A protective slime layer created by colonies of bacteria that resists cleaning and disinfectionExplanation: Biofilms form when bacteria attach to a surface and secrete a matrix (slime). They are very hard to remove and shield bacteria from chemicals.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q28.Robotic instruments (like DaVinci) typically require:
A.Less cleaning timeB.Standard washingC.Complex, validated cleaning protocols involving specific flushing and ultrasonic processingD.No cleaning✓C. Complex, validated cleaning protocols involving specific flushing and ultrasonic processingExplanation: Robotic arms have complex cables and pulleys. They require very specific, time-intensive manual and automated cleaning steps.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q29.The temperature of the decontamination area should be maintained between:
A.60°F - 65°F (16°C - 18°C)B.68°F - 73°F (20°C - 23°C)C.75°F - 80°F (24°C - 27°C)D.Whatever is comfortable✓A. 60°F - 65°F (16°C - 18°C)Explanation: Decon should be cooler (60-65°F) to inhibit bacterial growth and keep staff comfortable while wearing hot PPE.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q30.If instruments arrive from the OR with saline on them, what will likely happen?
A.They will be cleanerB.Pitting and corrosion will occurC.They will be sterileD.Nothing✓B. Pitting and corrosion will occurExplanation: Saline is salt water. Salt causes rapid corrosion (rust/pitting) on stainless steel. OR staff should use sterile water, not saline, to keep items moist.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q31.Degassing an ultrasonic cleaner is done to:
A.Remove dissolved gases from the water to improve cavitationB.Remove soilC.Heat the waterD.Cool the water✓A. Remove dissolved gases from the water to improve cavitationExplanation: Excess gas bubbles cushion the cavitation implosions, reducing cleaning power. Running the unit empty for a cycle degasses the water.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q32.Glutaraldehyde usually requires a soak time of ____ for high-level disinfection at 20°C.
A.5 minutesB.45-90 minutesC.10 hoursD.24 hours✓B. 45-90 minutesExplanation: For HLD, Glutaraldehyde typically requires 45-90 mins (depending on formulation). For Sterilization, it requires 10 hours.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q33.Phenolics are environmental disinfectants that should generally NOT be used on:
A.FloorsB.WallsC.Surgical instruments or items that contact skin/infantsD.Trash cans✓C. Surgical instruments or items that contact skin/infantsExplanation: Phenolics leave a residue that can irritate skin and is toxic to infants (depigmentation). They are for environmental surfaces only.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q34.The final step in manual cleaning is:
A.BrushingB.Enzymatic soakC.Thorough rinsing with critical waterD.Drying✓C. Thorough rinsing with critical waterExplanation: Rinsing removes the detergent and loosened soil. Critical water prevents recontamination/staining. (Drying follows, but rinsing is the final 'cleaning' step action).
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q35.Loaner instruments should be:
A.Sent straight to the ORB.Sterilized immediatelyC.Logged in, inspected, and completely reprocessed (cleaned and sterilized) by the facilityD.Wiped down only✓C. Logged in, inspected, and completely reprocessed (cleaned and sterilized) by the facilityExplanation: You cannot trust outside processing. Loaners must be treated as contaminated and go through the full decontamination cycle at the receiving facility.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q36.When cleaning a Kerrison Rongeur, the most critical step often missed is:
A.LubricationB.Testing sharpnessC.Disassembling the instrument or opening the slide mechanism to clean the trackD.Soaking it✓C. Disassembling the instrument or opening the slide mechanism to clean the trackExplanation: Kerrisons have a sliding mechanism that traps bone and tissue. If not disassembled or opened fully during cleaning, bioburden remains trapped inside, leading to sterilization failure.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q37.A Loaner Set arrives at 2:00 PM for a 3:00 PM case. The manufacturer IFU requires an 8-minute steam cycle. The hospital's policy allows IUSS. What is the BEST decision?
A.Flash (IUSS) the tray to get it there on timeB.Inform the OR that the case must be delayed to allow for proper reprocessingC.Wipe it down with alcohol and send it upD.Use the hospital's own instruments instead✓B. Inform the OR that the case must be delayed to allow for proper reprocessingExplanation: Patient safety comes first. Loaner trays must be fully decontaminated and sterilized. 1 hour is insufficient for proper processing (soak, clean, pack, sterilize, cool). Pressure to flash sterilize due to late delivery should be resisted.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q38.Which water quality issue creates a 'rainbow' effect or stain on stainless steel instruments?
A.High pHB.SilicatesC.ChlorideD.Iron✓B. SilicatesExplanation: Excessive silicates in the water supply typically cause blue/rainbow discoloration on instruments.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime
Q39.A 'TASS' outbreak in the ophthalmology department would trigger an audit of:
A.The steam qualityB.The cleaning and rinsing protocols for eye instruments (water quality, detergent residue)C.The air handling systemD.The peel pouches✓B. The cleaning and rinsing protocols for eye instruments (water quality, detergent residue)Explanation: Toxic Anterior Segment Syndrome (TASS) is caused by residues (detergents, endotoxins) left on eye instruments. The audit would focus on water quality (critical water) and rinsing steps.
See answer — start free trial3-day free trial · $9.99/mo after · cancel anytime