Sterile Processing Technician Exam
Decontamination Practice Questions
39 practice questions with detailed explanations — aligned to the Sterile Processing Technician Exam.
Q1.What is the first step in the decontamination process?
A.Manual cleaningB.Point-of-use treatmentC.SortingD.DisinfectionB. Point-of-use treatmentExplanation: Decontamination begins at the point-of-use (in the OR) to prevent blood/soil from drying on instruments. This makes subsequent cleaning steps effective.
Join 1,000+ users passing the Sterile Processing Technician
Q2.The ideal temperature for enzymatic cleaners is typically:
A.Above 140°F (60°C)B.Below 100°F (38°C)C.Between 109°F - 140°F (43°C - 60°C)D.BoilingC. Between 109°F - 140°F (43°C - 60°C)Explanation: Enzymes are proteins that work best at specific warm temperatures. Too hot (>140°F) kills the enzymes; too cold renders them sluggish. Always check manufacturer IFUs.
Join 1,000+ users passing the Sterile Processing Technician
Q3.Cavitation is the cleaning mechanism used in:
A.Washer-disinfectorsB.Ultrasonic cleanersC.Cart washersD.Steam sterilizersB. Ultrasonic cleanersExplanation: Ultrasonic cleaners use sound waves to create microscopic bubbles that implode (cavitation), pulling soil off surfaces and crevices.
Join 1,000+ users passing the Sterile Processing Technician
Q4.Which type of water is recommended for the final rinse of critical medical devices?
A.Tap waterB.Critical water (Deionized/Distilled/Reverse Osmosis)C.Softened waterD.Utility waterB. Critical water (Deionized/Distilled/Reverse Osmosis)Explanation: Final rinses require water free of minerals and impurities (Critical Water) to prevent staining, corrosion, and pyrogen contamination.
Join 1,000+ users passing the Sterile Processing Technician
Q5.When manually cleaning a device, the brushing motion should be performed:
A.Above the water line to see clearlyB.Under the water line to prevent aerosolizationC.Vigorously with a wire brushD.Only on the exterior surfacesB. Under the water line to prevent aerosolizationExplanation: Brushing under the water surface prevents aerosols containing pathogens from becoming airborne and contaminating the technician.
Join 1,000+ users passing the Sterile Processing Technician
Q6.A washer-disinfector uses which process to achieve thermal disinfection?
A.Chemical exposureB.ImpingementC.Hot water rinse (typically 180°F - 203°F)D.Ultrasonic wavesC. Hot water rinse (typically 180°F - 203°F)Explanation: Thermal disinfection in automated washers relies on sustaining a high temperature rinse for a specific time to kill vegetative microorganisms.
Join 1,000+ users passing the Sterile Processing Technician
Q7.Items with lumens (channels) must be:
A.Soaked onlyB.Brushed with a properly sized brush and flushedC.Processed in the ultrasonic onlyD.Sterilized without cleaningB. Brushed with a properly sized brush and flushedExplanation: Soaking isn't enough. Friction (brushing) is required to remove soil inside the lumen, followed by flushing to clear debris.
Join 1,000+ users passing the Sterile Processing Technician
Q8.The pH of a detergent used for anodized aluminum containers (rigid containers) should be:
A.High alkalineB.Neutral (pH 7)C.AcidicD.Low alkalineB. Neutral (pH 7)Explanation: Anodized aluminum is damaged by high/low pH. Neutral pH detergents are required to prevent corrosion/pitting of the protective layer.
Join 1,000+ users passing the Sterile Processing Technician
Q9.Proper airflow in the decontamination area should be:
A.Positive pressureB.Negative pressureC.Neutral pressureD.It doesn't matterB. Negative pressureExplanation: Negative pressure ensures that air flows *into* the decon area from clean areas and is exhausted outside, preventing contaminants from drifting into clean prep areas.
Join 1,000+ users passing the Sterile Processing Technician
Q10.TASS (Toxic Anterior Segment Syndrome) is an acute inflammation of the eye caused by:
A.BacteriaB.Inadequate removal of detergent or chemical residues from ophthalmic instrumentsC.Viral infectionD.Patient allergiesB. Inadequate removal of detergent or chemical residues from ophthalmic instrumentsExplanation: TASS is a sterile inflammatory reaction caused by toxic residues (like detergents, endotoxins, or enzymes) left on eye instruments. Thorough rinsing with critical water is essential.
Join 1,000+ users passing the Sterile Processing Technician
Q11.Before placing instruments into an ultrasonic cleaner, they must be:
A.SterilizedB.Grossly cleaned of visible soilC.DryD.LubricatedB. Grossly cleaned of visible soilExplanation: Ultrasonic cleaners are for fine cleaning. Gross soil inhibits cavitation and contaminates the bath quickly. Pre-cleaning is mandatory.
Join 1,000+ users passing the Sterile Processing Technician
Q12.Descaling is the process of:
A.Removing instruments from a trayB.Removing scale (mineral deposits) from the chamber of a washer or sterilizerC.Checking weightD.Lubricating instrumentsB. Removing scale (mineral deposits) from the chamber of a washer or sterilizerExplanation: Hard water leaves chalky mineral deposits (scale) on equipment walls. Descaling agents (acidic) remove this buildup to maintain efficiency.
Join 1,000+ users passing the Sterile Processing Technician
Q13.Which agency regulates disinfectants and sterilants used on non-living surfaces?
A.FDAB.EPA (Environmental Protection Agency)C.OSHAD.CDCB. EPA (Environmental Protection Agency)Explanation: The EPA regulates surface disinfectants and liquid chemical sterilants (as pesticides). The FDA regulates high-level disinfectants used on medical devices.
Join 1,000+ users passing the Sterile Processing Technician
Q14.Personal Protective Equipment (PPE) for the decontamination area includes:
A.Lab coat onlyB.Gloves, fluid-resistant gown, eye protection, face mask/shield, and shoe coversC.Sterile gloves and gownD.Street clothesB. Gloves, fluid-resistant gown, eye protection, face mask/shield, and shoe coversExplanation: Full PPE is required to protect against splashes and biohazards. Heavy-duty utility gloves are preferred over exam gloves.
Join 1,000+ users passing the Sterile Processing Technician
Q15.A 'High-Level Disinfectant' (HLD) kills:
A.All microorganisms including high numbers of bacterial sporesB.All vegetative bacteria, viruses, fungi, and mycobacteria, but not necessarily high numbers of sporesC.Only bacteriaD.Only virusesB. All vegetative bacteria, viruses, fungi, and mycobacteria, but not necessarily high numbers of sporesExplanation: HLD kills almost everything (TB, viruses, fungi) but cannot guarantee the kill of high numbers of spores (that requires sterilization).
Join 1,000+ users passing the Sterile Processing Technician
Q16.The minimum effective concentration (MEC) of a high-level disinfectant must be tested:
A.Once a weekB.Once a dayC.Before each use (or load)D.When the bottle is openedC. Before each use (or load)Explanation: HLD solutions (like Glutaraldehyde or OPA) degrade over time. You must dip a test strip before *every* use to ensure potency.
Join 1,000+ users passing the Sterile Processing Technician
Q17.According to the Spaulding Classification, 'Semi-critical' items require at least:
A.Cleaning onlyB.Low-level disinfectionC.High-level disinfectionD.SterilizationC. High-level disinfectionExplanation: Semi-critical items touch mucous membranes (e.g., endoscopes, anesthesia blades) and require High-Level Disinfection. Critical items (enter sterile tissue) require sterilization.
Join 1,000+ users passing the Sterile Processing Technician
Q18.Why must instruments be disassembled before cleaning?
A.To count the partsB.To ensure cleaning agents contact all surfacesC.To make them fit in the trayD.It is not requiredB. To ensure cleaning agents contact all surfacesExplanation: Bioburden hides under movable parts. Disassembly ensures detergent and water reach every surface.
Join 1,000+ users passing the Sterile Processing Technician
Q19.Ortho-phthalaldehyde (OPA) is a common:
A.Enzymatic cleanerB.High-level disinfectantC.Sterilant gasD.LubricantB. High-level disinfectantExplanation: OPA is a widely used liquid high-level disinfectant for flexible endoscopes. It is a safer alternative to glutaraldehyde.
Join 1,000+ users passing the Sterile Processing Technician
Q20.If a rigid container is returned to decontamination with the filter still in place:
A.The filter can be reused if cleanB.The filter must be removed and discarded, and the retention plate cleanedC.The container can be washed with the filterD.Leave it for the assembly sideB. The filter must be removed and discarded, and the retention plate cleanedExplanation: Disposable filters are single-use. They must be removed in decon to allow proper cleaning of the lid and retention plate.
Join 1,000+ users passing the Sterile Processing Technician
Q21.When loading a mechanical washer, instruments should be placed:
A.Stacked on top of each otherB.In the closed positionC.In the open/unlocked positionD.In peel pouchesC. In the open/unlocked positionExplanation: Box locks and ratchets must be open to allow spray action to reach the hinge areas and jaws.
Join 1,000+ users passing the Sterile Processing Technician
Q22.The 'hollow' cavitation test used for ultrasonic cleaners involves:
A.A ceramic discB.Aluminum foilC.A TOSI testD.A biological indicatorB. Aluminum foilExplanation: The aluminum foil test detects uneven cavitation. (Note: Newer commercial tests like SonoCheck are also used, but foil is the classic textbook answer for verifying cavitation energy distribution).
Join 1,000+ users passing the Sterile Processing Technician
Q23.Prions (which cause CJD) are uniquely resistant to:
A.Standard cleaningB.Standard sterilization cycles (normal steam/EtO)C.High-level disinfectionD.All of the aboveD. All of the aboveExplanation: Prions are misfolded proteins that are incredibly resistant. They require extended sterilization cycles (e.g., 18 min at 273°F or 1 hour at 250°F) and specific cleaning protocols.
Join 1,000+ users passing the Sterile Processing Technician
Q24.Which detergent is best for removing stubborn organic soils like dried blood?
A.Acidic detergentB.Enzymatic detergent (Protease)C.High alkaline detergentD.Neutral detergentB. Enzymatic detergent (Protease)Explanation: Protease enzymes specifically break down proteins (blood/tissue). Lipase breaks down fats.
Join 1,000+ users passing the Sterile Processing Technician
Q25.Transporting contaminated items to the decontamination area requires:
A.Carrying them in handB.A closed, leak-proof container marked with a biohazard labelC.An open wire cartD.A clear plastic bagB. A closed, leak-proof container marked with a biohazard labelExplanation: OSHA requires biohazardous materials to be contained in leak-proof, labeled containers to protect staff and the environment during transport.
Join 1,000+ users passing the Sterile Processing Technician
Q26.Power equipment (drills, saws) should generally be:
A.Immersed in waterB.Cleaned manually (wiped down) and never immersedC.Processed in the ultrasonicD.Put in the washer-disinfectorB. Cleaned manually (wiped down) and never immersedExplanation: Immersion allows water to enter the motor housing, destroying the mechanism. Follow IFU carefully—usually manual wipe-down.
Join 1,000+ users passing the Sterile Processing Technician
Q27.Biofilms are:
A.A type of soapB.A protective slime layer created by colonies of bacteria that resists cleaning and disinfectionC.A sterile barrierD.A lubricantB. A protective slime layer created by colonies of bacteria that resists cleaning and disinfectionExplanation: Biofilms form when bacteria attach to a surface and secrete a matrix (slime). They are very hard to remove and shield bacteria from chemicals.
Join 1,000+ users passing the Sterile Processing Technician
Q28.Robotic instruments (like DaVinci) typically require:
A.Less cleaning timeB.Standard washingC.Complex, validated cleaning protocols involving specific flushing and ultrasonic processingD.No cleaningC. Complex, validated cleaning protocols involving specific flushing and ultrasonic processingExplanation: Robotic arms have complex cables and pulleys. They require very specific, time-intensive manual and automated cleaning steps.
Join 1,000+ users passing the Sterile Processing Technician
Q29.The temperature of the decontamination area should be maintained between:
A.60°F - 65°F (16°C - 18°C)B.68°F - 73°F (20°C - 23°C)C.75°F - 80°F (24°C - 27°C)D.Whatever is comfortableA. 60°F - 65°F (16°C - 18°C)Explanation: Decon should be cooler (60-65°F) to inhibit bacterial growth and keep staff comfortable while wearing hot PPE.
Join 1,000+ users passing the Sterile Processing Technician
Q30.If instruments arrive from the OR with saline on them, what will likely happen?
A.They will be cleanerB.Pitting and corrosion will occurC.They will be sterileD.NothingB. Pitting and corrosion will occurExplanation: Saline is salt water. Salt causes rapid corrosion (rust/pitting) on stainless steel. OR staff should use sterile water, not saline, to keep items moist.
Join 1,000+ users passing the Sterile Processing Technician
Q31.Degassing an ultrasonic cleaner is done to:
A.Remove dissolved gases from the water to improve cavitationB.Remove soilC.Heat the waterD.Cool the waterA. Remove dissolved gases from the water to improve cavitationExplanation: Excess gas bubbles cushion the cavitation implosions, reducing cleaning power. Running the unit empty for a cycle degasses the water.
Join 1,000+ users passing the Sterile Processing Technician
Q32.Glutaraldehyde usually requires a soak time of ____ for high-level disinfection at 20°C.
A.5 minutesB.45-90 minutesC.10 hoursD.24 hoursB. 45-90 minutesExplanation: For HLD, Glutaraldehyde typically requires 45-90 mins (depending on formulation). For Sterilization, it requires 10 hours.
Join 1,000+ users passing the Sterile Processing Technician
Q33.Phenolics are environmental disinfectants that should generally NOT be used on:
A.FloorsB.WallsC.Surgical instruments or items that contact skin/infantsD.Trash cansC. Surgical instruments or items that contact skin/infantsExplanation: Phenolics leave a residue that can irritate skin and is toxic to infants (depigmentation). They are for environmental surfaces only.
Join 1,000+ users passing the Sterile Processing Technician
Q34.The final step in manual cleaning is:
A.BrushingB.Enzymatic soakC.Thorough rinsing with critical waterD.DryingC. Thorough rinsing with critical waterExplanation: Rinsing removes the detergent and loosened soil. Critical water prevents recontamination/staining. (Drying follows, but rinsing is the final 'cleaning' step action).
Join 1,000+ users passing the Sterile Processing Technician
Q35.Loaner instruments should be:
A.Sent straight to the ORB.Sterilized immediatelyC.Logged in, inspected, and completely reprocessed (cleaned and sterilized) by the facilityD.Wiped down onlyC. Logged in, inspected, and completely reprocessed (cleaned and sterilized) by the facilityExplanation: You cannot trust outside processing. Loaners must be treated as contaminated and go through the full decontamination cycle at the receiving facility.
Join 1,000+ users passing the Sterile Processing Technician
Q36.When cleaning a Kerrison Rongeur, the most critical step often missed is:
A.LubricationB.Testing sharpnessC.Disassembling the instrument or opening the slide mechanism to clean the trackD.Soaking itC. Disassembling the instrument or opening the slide mechanism to clean the trackExplanation: Kerrisons have a sliding mechanism that traps bone and tissue. If not disassembled or opened fully during cleaning, bioburden remains trapped inside, leading to sterilization failure.
Join 1,000+ users passing the Sterile Processing Technician
Q37.A Loaner Set arrives at 2:00 PM for a 3:00 PM case. The manufacturer IFU requires an 8-minute steam cycle. The hospital's policy allows IUSS. What is the BEST decision?
A.Flash (IUSS) the tray to get it there on timeB.Inform the OR that the case must be delayed to allow for proper reprocessingC.Wipe it down with alcohol and send it upD.Use the hospital's own instruments insteadB. Inform the OR that the case must be delayed to allow for proper reprocessingExplanation: Patient safety comes first. Loaner trays must be fully decontaminated and sterilized. 1 hour is insufficient for proper processing (soak, clean, pack, sterilize, cool). Pressure to flash sterilize due to late delivery should be resisted.
Join 1,000+ users passing the Sterile Processing Technician
Q38.Which water quality issue creates a 'rainbow' effect or stain on stainless steel instruments?
A.High pHB.SilicatesC.ChlorideD.IronB. SilicatesExplanation: Excessive silicates in the water supply typically cause blue/rainbow discoloration on instruments.
Join 1,000+ users passing the Sterile Processing Technician
Q39.A 'TASS' outbreak in the ophthalmology department would trigger an audit of:
A.The steam qualityB.The cleaning and rinsing protocols for eye instruments (water quality, detergent residue)C.The air handling systemD.The peel pouchesB. The cleaning and rinsing protocols for eye instruments (water quality, detergent residue)Explanation: Toxic Anterior Segment Syndrome (TASS) is caused by residues (detergents, endotoxins) left on eye instruments. The audit would focus on water quality (critical water) and rinsing steps.