38 practice questions with detailed explanations — aligned to the Sterile Processing Technician Exam.

  1. Q1.Which biological indicator (spore) is used to test Steam sterilization cycles?

    A.Bacillus atrophaeus
    B.Geobacillus stearothermophilus
    C.Bacillus subtilis
    D.Clostridium tetani
    BGeobacillus stearothermophilus

    Explanation: Geobacillus stearothermophilus is the spore most resistant to steam sterilization. If these spores are killed, the cycle is considered effective.

    Join 1,000+ users passing the Sterile Processing Technician

  2. Q2.Which biological indicator is used for Ethylene Oxide (EtO) and Dry Heat sterilization?

    A.Geobacillus stearothermophilus
    B.Bacillus atrophaeus
    C.Geobacillus thermoglucosidasius
    D.Staphylococcus aureus
    BBacillus atrophaeus

    Explanation: Bacillus atrophaeus is the specific spore used to challenge EtO and Dry Heat cycles due to its high resistance to dry/chemical sterilization.

    Join 1,000+ users passing the Sterile Processing Technician

  3. Q3.A Bowie-Dick test is designed to check for:

    A.Sterility of the load
    B.Proper temperature
    C.Air removal efficiency and air leaks in dynamic air removal (prevacuum) sterilizers
    D.Water quality
    CAir removal efficiency and air leaks in dynamic air removal (prevacuum) sterilizers

    Explanation: The Bowie-Dick test is a Class 2 Chemical Indicator used daily to verify that the vacuum system effectively removes air from the chamber. It does NOT prove sterilization.

    Join 1,000+ users passing the Sterile Processing Technician

  4. Q4.When should the Bowie-Dick test be run?

    A.At the end of the day
    B.Each day the sterilizer is used, before the first processed load, in an empty chamber
    C.With the first full load of instruments
    D.Once a week
    BEach day the sterilizer is used, before the first processed load, in an empty chamber

    Explanation: The test must be run in an empty chamber (to maximize the challenge to the vacuum system) at the beginning of each day of use.

    Join 1,000+ users passing the Sterile Processing Technician

  5. Q5.A Class 5 Chemical Integrator:

    A.Reacts to only one parameter (temperature)
    B.Is used only on the outside of packages
    C.Reacts to all critical variables of the sterilization cycle and correlates to the biological kill
    D.Is a Bowie-Dick test
    CReacts to all critical variables of the sterilization cycle and correlates to the biological kill

    Explanation: Class 5 Integrators are internal indicators designed to mimic the response of a biological indicator by reacting to time, temperature, and steam quality.

    Join 1,000+ users passing the Sterile Processing Technician

  6. Q6.Where should the Process Challenge Device (PCD) containing the Biological Indicator be placed in a steam sterilizer?

    A.Top shelf, back
    B.Center of the load
    C.Bottom shelf, directly over the drain
    D.Top shelf, front
    CBottom shelf, directly over the drain

    Explanation: The area over the drain is the 'coldest point' in the sterilizer because air and condensate gravitate there. Placing the PCD here presents the greatest challenge to the cycle.

    Join 1,000+ users passing the Sterile Processing Technician

  7. Q7.Which Class of FDA Product Recall is the most serious, indicating a reasonable probability of serious health problems or death?

    A.Class I
    B.Class II
    C.Class III
    D.Class IV
    AClass I

    Explanation: Class I recalls involve products that could cause serious injury or death. Class II involves temporary/reversible health issues. Class III is unlikely to cause adverse health consequences.

    Join 1,000+ users passing the Sterile Processing Technician

  8. Q8.Validation vs. Verification: Who performs 'Validation' of a sterilization process?

    A.The end-user (Technician)
    B.The device manufacturer
    C.The FDA
    D.The hospital administration
    BThe device manufacturer

    Explanation: Validation is performed by the manufacturer to prove their instructions work. Verification is performed by the end-user (CS tech) to confirm they are following those instructions correctly.

    Join 1,000+ users passing the Sterile Processing Technician

  9. Q9.According to AAMI ST79, how often should a Biological Indicator (BI) test be run for steam sterilizers?

    A.Monthly
    B.At least weekly, but preferably daily and with every load containing implants
    C.Every other day
    D.Only when the machine breaks
    BAt least weekly, but preferably daily and with every load containing implants

    Explanation: Weekly is the absolute minimum requirement, but daily (plus every implant load) is the recommended best practice for patient safety.

    Join 1,000+ users passing the Sterile Processing Technician

  10. Q10.If a biological indicator test is positive, what is the first action the technician should take?

    A.Run another cycle immediately
    B.Report it to the supervisor and recall all loads processed since the last negative BI
    C.Ignore it if the chemical integrator passed
    D.Incubate it for another 24 hours
    BReport it to the supervisor and recall all loads processed since the last negative BI

    Explanation: A positive BI means sterilization failed (spores grew). All items processed since the last confirmed 'good' cycle are suspect and must be recalled/quarantined.

    Join 1,000+ users passing the Sterile Processing Technician

  11. Q11.Which Chemical Indicator is used to differentiate between processed and unprocessed items (external indicator)?

    A.Class 1
    B.Class 3
    C.Class 5
    D.Class 6
    AClass 1

    Explanation: Class 1 Process Indicators (like autoclave tape) change color to show the item has been exposed to the sterilization process, preventing unprocessed items from being used.

    Join 1,000+ users passing the Sterile Processing Technician

  12. Q12.What information does a 'Load Control Label' provide?

    A.The sterilization method, date, sterilizer number, and load number
    B.The patient's name and room number
    C.The cost of the instrument set
    D.The manufacturer's phone number
    AThe sterilization method, date, sterilizer number, and load number

    Explanation: Load labels (Lot numbers) provide the traceability needed to recall specific items if a sterilization failure occurs.

    Join 1,000+ users passing the Sterile Processing Technician

  13. Q13.When incubating a Biological Indicator, what is the purpose of the 'Control' vial?

    A.To calibrate the incubator
    B.To prove the spores were alive/viable before the cycle and the incubator is functioning correctly
    C.To compare color change
    D.To test the sterilizer
    BTo prove the spores were alive/viable before the cycle and the incubator is functioning correctly

    Explanation: The control vial is not sterilized. It MUST grow bacteria (turn yellow/positive). If the control doesn't grow, the test is invalid because the spores might have been dead to begin with.

    Join 1,000+ users passing the Sterile Processing Technician

  14. Q14.Implantable devices should not be released for use until:

    A.The cycle is finished
    B.The physical monitor printout is checked
    C.The Biological Indicator result is known (negative)
    D.The surgeon calls for it
    CThe Biological Indicator result is known (negative)

    Explanation: Because implants remain in the body, the risk of infection is higher. Loads containing implants should be quarantined until the BI confirms sterilization success.

    Join 1,000+ users passing the Sterile Processing Technician

  15. Q15.A Class 6 'Emulating Indicator' verifies:

    A.That the package is sterile
    B.That all critical parameters for a *specific* cycle (e.g., 270°F for 4 min) were met
    C.External exposure only
    D.Air removal only
    BThat all critical parameters for a *specific* cycle (e.g., 270°F for 4 min) were met

    Explanation: Class 6 indicators are cycle-specific. They emulate the performance of the specific cycle they are designed for (temperature/time) with a high degree of precision.

    Join 1,000+ users passing the Sterile Processing Technician

  16. Q16.Testing the sharpness of scissors or the alignment of hemostats is a form of:

    A.Sterilization
    B.Decontamination
    C.Quality Assurance / Inspection
    D.Packaging
    CQuality Assurance / Inspection

    Explanation: Inspection and functional testing ensure the device will perform as intended. Sending up a broken instrument is a quality failure.

    Join 1,000+ users passing the Sterile Processing Technician

  17. Q17.Which organization develops the nationally recognized standards (like ST79) for sterilization?

    A.OSHA
    B.FDA
    C.AAMI (Association for the Advancement of Medical Instrumentation)
    D.EPA
    CAAMI (Association for the Advancement of Medical Instrumentation)

    Explanation: AAMI sets the consensus standards for sterilization processing. OSHA regulates safety; FDA regulates device manufacturing/sales.

    Join 1,000+ users passing the Sterile Processing Technician

  18. Q18.Temperature and humidity logs in the sterile storage area should be checked and recorded:

    A.Hourly
    B.Daily
    C.Weekly
    D.Monthly
    BDaily

    Explanation: Daily monitoring ensures environmental conditions remain within safe limits to maintain sterility.

    Join 1,000+ users passing the Sterile Processing Technician

  19. Q19.During a recall, the first step is to:

    A.Call the patients
    B.Identify all items processed in the failed load using lot control numbers
    C.Throw away the sterilizer
    D.Re-sterilize everything in the department
    BIdentify all items processed in the failed load using lot control numbers

    Explanation: Traceability is key. Use the lot numbers to find exactly which items were in the failed load and retrieve them.

    Join 1,000+ users passing the Sterile Processing Technician

  20. Q20.Which test is used to verify the efficacy of the mechanical washer's cleaning ability (cavitation or impingement)?

    A.Bowie-Dick Test
    B.TOSI (Test Object Surgical Instrument) or foil test
    C.Geobacillus stearothermophilus
    D.Class 5 Integrator
    BTOSI (Test Object Surgical Instrument) or foil test

    Explanation: Cleaning verification tests like TOSI (dried blood soil simulation) or aluminum foil tests verify that the washing equipment is actually removing soil.

    Join 1,000+ users passing the Sterile Processing Technician

  21. Q21.Sterilization records (printouts, BI results) must be retained for how long?

    A.1 year
    B.As specified by local/state statutes or facility policy (often 3+ years)
    C.Until the patient is discharged
    D.Forever
    BAs specified by local/state statutes or facility policy (often 3+ years)

    Explanation: Retention requirements vary by state, but records are typically kept for several years to align with statutes of limitations for lawsuits.

    Join 1,000+ users passing the Sterile Processing Technician

  22. Q22.If the physical monitor (printer) on the sterilizer indicates a failure (e.g., temperature dropped), but the chemical indicator passed, what should be done?

    A.Release the load based on the CI
    B.Consider the load unsterile and reprocess
    C.Ask the surgeon
    D.Run a BI test
    BConsider the load unsterile and reprocess

    Explanation: The physical monitor is the first and most critical indicator. If the printout shows parameters were not met, the load is unsterile, regardless of what the CIs say.

    Join 1,000+ users passing the Sterile Processing Technician

  23. Q23.Using a rigid container that has not been validated for use in your specific sterilizer cycle is an example of:

    A.Off-label use
    B.Validation
    C.Verification
    D.Quality control
    AOff-label use

    Explanation: Using a device in a manner not approved by the manufacturer (IFU) is 'off-label' and creates liability/safety risks.

    Join 1,000+ users passing the Sterile Processing Technician

  24. Q24.Which of the following is considered a 'Sentinel Event' in sterilization?

    A.A torn wrapper found in storage
    B.A biological indicator turning yellow (positive)
    C.Sending a non-sterile instrument to the OR that is used on a patient
    D.The sterilizer running out of paper
    CSending a non-sterile instrument to the OR that is used on a patient

    Explanation: A sentinel event is an unexpected occurrence involving death or serious physical injury. Using non-sterile instruments on a patient fits this criteria (Joint Commission).

    Join 1,000+ users passing the Sterile Processing Technician

  25. Q25.The recommended incubation temperature for Geobacillus stearothermophilus is:

    A.35°C - 39°C (95°F - 102°F)
    B.55°C - 60°C (131°F - 140°F)
    C.98.6°F (37°C)
    D.Room temperature
    B55°C - 60°C (131°F - 140°F)

    Explanation: Geobacillus stearothermophilus is a thermophile (heat-loving). It requires high incubation temperatures (55-60°C) to grow.

    Join 1,000+ users passing the Sterile Processing Technician

  26. Q26.A chemical indicator placed inside a rigid container is called an:

    A.External indicator
    B.Internal indicator
    C.Bowie-Dick
    D.Equipment monitor
    BInternal indicator

    Explanation: Internal indicators verify that the sterilant penetrated the barrier (container) and reached the instruments inside.

    Join 1,000+ users passing the Sterile Processing Technician

  27. Q27.If wet packs are frequently found in a specific sterilizer, the first step in troubleshooting should be:

    A.Call the repair technician immediately
    B.Review loading procedures (metal mass, spacing, tilting)
    C.Increase drying time to 2 hours
    D.Use different wrappers
    BReview loading procedures (metal mass, spacing, tilting)

    Explanation: Most wet packs are caused by operator error (improper loading). Verify procedures before assuming equipment failure.

    Join 1,000+ users passing the Sterile Processing Technician

  28. Q28.Which agency requires Safety Data Sheets (SDS) be available for all chemicals used in the department?

    A.CDC
    B.OSHA
    C.FDA
    D.AAMI
    BOSHA

    Explanation: OSHA's Hazard Communication Standard ('Right to Know') mandates access to SDS for employee safety.

    Join 1,000+ users passing the Sterile Processing Technician

  29. Q29.Product quality assurance testing of a rigid sterilization container includes checking the:

    A.Color of the lid
    B.Integrity of the gasket and retention plate/filter mechanisms
    C.Weight of the empty container
    D.Brand name
    BIntegrity of the gasket and retention plate/filter mechanisms

    Explanation: Routine inspection must verify the gasket seals tight and the filter retention mechanism is not bent or loose, ensuring a sterile barrier.

    Join 1,000+ users passing the Sterile Processing Technician

  30. Q30.For quality assurance, a Process Challenge Device (PCD) for EtO sterilization should contain:

    A.A Bowie-Dick test
    B.A biological indicator (Bacillus atrophaeus) and a chemical indicator
    C.A thermometer
    D.A syringe
    BA biological indicator (Bacillus atrophaeus) and a chemical indicator

    Explanation: EtO PCDs typically contain a BI (B. atrophaeus), a CI, and materials (like a syringe or tubing) that create a resistance challenge similar to the load.

    Join 1,000+ users passing the Sterile Processing Technician

  31. Q31.Failure to remove soil from the box lock of a hemostat during cleaning is a failure of:

    A.Sterilization
    B.Decontamination
    C.Packaging
    D.Transport
    BDecontamination

    Explanation: Cleaning is the core of decontamination. If soil remains, sterility cannot be achieved.

    Join 1,000+ users passing the Sterile Processing Technician

  32. Q32.A rapid-readout Biological Indicator uses what technology to detect spores?

    A.Visual color change (pH shift) only
    B.Fluorescence (enzyme detection)
    C.Smell
    D.Microscope
    BFluorescence (enzyme detection)

    Explanation: Rapid-readout BIs (1-3 hours) detect fluorescence produced by enzymes released by living spores, providing results much faster than visual pH color change (24-48 hours).

    Join 1,000+ users passing the Sterile Processing Technician

  33. Q33.What is the correct placement for a Bowie-Dick test pack?

    A.In a full load, bottom shelf
    B.In an empty chamber, bottom shelf, over the drain
    C.In an empty chamber, top shelf
    D.In a full load, center
    BIn an empty chamber, bottom shelf, over the drain

    Explanation: The test pack must be placed in the most challenging area (over the drain) in an empty chamber to accurately test the vacuum system's ability to remove air.

    Join 1,000+ users passing the Sterile Processing Technician

  34. Q34.Verification of the ultrasonic cleaner's cavitation energy should be performed:

    A.Annually
    B.At least weekly (or preferably daily)
    C.Never
    D.When it makes a loud noise
    BAt least weekly (or preferably daily)

    Explanation: Routine verification (weekly/daily) ensures the transducers are functioning and cleaning is effective.

    Join 1,000+ users passing the Sterile Processing Technician

  35. Q35.Which document is the ultimate authority on how to process a specific medical device?

    A.AAMI ST79
    B.The textbook
    C.The Device Manufacturer's Instructions for Use (IFU)
    D.The Manager's opinion
    CThe Device Manufacturer's Instructions for Use (IFU)

    Explanation: The Manufacturer's IFU is the primary source of truth. Validated instructions must be followed to ensure safety and efficacy.

    Join 1,000+ users passing the Sterile Processing Technician

  36. Q36.A load record indicates that a 'Control' Biological Indicator was NOT run with the daily BI test. The test BI came back negative (growth). What is the status of the sterilizer?

    A.The sterilizer is functioning correctly
    B.The test is invalid and inconclusive
    C.The sterilizer failed
    D.The load must be recalled
    BThe test is invalid and inconclusive

    Explanation: Without a positive Control BI, you cannot prove the spores were alive to begin with or that the incubator is working. A negative test result is meaningless (invalid) without a positive control for comparison.

    Join 1,000+ users passing the Sterile Processing Technician

  37. Q37.Which scenario represents a 'sentinel event'?

    A.A sterilizer breaks down
    B.A patient develops a deep surgical site infection resulting in loss of limb due to non-sterile instruments
    C.A technician is late
    D.A wrapper is torn
    BA patient develops a deep surgical site infection resulting in loss of limb due to non-sterile instruments

    Explanation: A sentinel event is an unexpected occurrence involving death or serious physical/psychological injury. A severe infection from dirty instruments qualifies.

    Join 1,000+ users passing the Sterile Processing Technician

  38. Q38.When testing a rigid sterilization container, a technician places a biological indicator inside but forgets the chemical indicator. The BI is negative (pass). Can the tray be used?

    A.Yes, the BI passed
    B.No, a chemical indicator is required inside every package to verify exposure
    C.Yes, if the external lock is intact
    D.No, the container must be thrown away
    BNo, a chemical indicator is required inside every package to verify exposure

    Explanation: Every package must contain an internal chemical indicator (Class 3, 4, 5, or 6) to verify that sterilant penetrated the package. Without it, the package is considered non-compliant/unsterile.

    Join 1,000+ users passing the Sterile Processing Technician