Sterile Processing Technician Exam
Quality Assurance Practice Questions
38 practice questions with detailed explanations — aligned to the Sterile Processing Technician Exam.
Q1.Which biological indicator (spore) is used to test Steam sterilization cycles?
A.Bacillus atrophaeusB.Geobacillus stearothermophilusC.Bacillus subtilisD.Clostridium tetaniB. Geobacillus stearothermophilusExplanation: Geobacillus stearothermophilus is the spore most resistant to steam sterilization. If these spores are killed, the cycle is considered effective.
Join 1,000+ users passing the Sterile Processing Technician
Q2.Which biological indicator is used for Ethylene Oxide (EtO) and Dry Heat sterilization?
A.Geobacillus stearothermophilusB.Bacillus atrophaeusC.Geobacillus thermoglucosidasiusD.Staphylococcus aureusB. Bacillus atrophaeusExplanation: Bacillus atrophaeus is the specific spore used to challenge EtO and Dry Heat cycles due to its high resistance to dry/chemical sterilization.
Join 1,000+ users passing the Sterile Processing Technician
Q3.A Bowie-Dick test is designed to check for:
A.Sterility of the loadB.Proper temperatureC.Air removal efficiency and air leaks in dynamic air removal (prevacuum) sterilizersD.Water qualityC. Air removal efficiency and air leaks in dynamic air removal (prevacuum) sterilizersExplanation: The Bowie-Dick test is a Class 2 Chemical Indicator used daily to verify that the vacuum system effectively removes air from the chamber. It does NOT prove sterilization.
Join 1,000+ users passing the Sterile Processing Technician
Q4.When should the Bowie-Dick test be run?
A.At the end of the dayB.Each day the sterilizer is used, before the first processed load, in an empty chamberC.With the first full load of instrumentsD.Once a weekB. Each day the sterilizer is used, before the first processed load, in an empty chamberExplanation: The test must be run in an empty chamber (to maximize the challenge to the vacuum system) at the beginning of each day of use.
Join 1,000+ users passing the Sterile Processing Technician
Q5.A Class 5 Chemical Integrator:
A.Reacts to only one parameter (temperature)B.Is used only on the outside of packagesC.Reacts to all critical variables of the sterilization cycle and correlates to the biological killD.Is a Bowie-Dick testC. Reacts to all critical variables of the sterilization cycle and correlates to the biological killExplanation: Class 5 Integrators are internal indicators designed to mimic the response of a biological indicator by reacting to time, temperature, and steam quality.
Join 1,000+ users passing the Sterile Processing Technician
Q6.Where should the Process Challenge Device (PCD) containing the Biological Indicator be placed in a steam sterilizer?
A.Top shelf, backB.Center of the loadC.Bottom shelf, directly over the drainD.Top shelf, frontC. Bottom shelf, directly over the drainExplanation: The area over the drain is the 'coldest point' in the sterilizer because air and condensate gravitate there. Placing the PCD here presents the greatest challenge to the cycle.
Join 1,000+ users passing the Sterile Processing Technician
Q7.Which Class of FDA Product Recall is the most serious, indicating a reasonable probability of serious health problems or death?
A.Class IB.Class IIC.Class IIID.Class IVA. Class IExplanation: Class I recalls involve products that could cause serious injury or death. Class II involves temporary/reversible health issues. Class III is unlikely to cause adverse health consequences.
Join 1,000+ users passing the Sterile Processing Technician
Q8.Validation vs. Verification: Who performs 'Validation' of a sterilization process?
A.The end-user (Technician)B.The device manufacturerC.The FDAD.The hospital administrationB. The device manufacturerExplanation: Validation is performed by the manufacturer to prove their instructions work. Verification is performed by the end-user (CS tech) to confirm they are following those instructions correctly.
Join 1,000+ users passing the Sterile Processing Technician
Q9.According to AAMI ST79, how often should a Biological Indicator (BI) test be run for steam sterilizers?
A.MonthlyB.At least weekly, but preferably daily and with every load containing implantsC.Every other dayD.Only when the machine breaksB. At least weekly, but preferably daily and with every load containing implantsExplanation: Weekly is the absolute minimum requirement, but daily (plus every implant load) is the recommended best practice for patient safety.
Join 1,000+ users passing the Sterile Processing Technician
Q10.If a biological indicator test is positive, what is the first action the technician should take?
A.Run another cycle immediatelyB.Report it to the supervisor and recall all loads processed since the last negative BIC.Ignore it if the chemical integrator passedD.Incubate it for another 24 hoursB. Report it to the supervisor and recall all loads processed since the last negative BIExplanation: A positive BI means sterilization failed (spores grew). All items processed since the last confirmed 'good' cycle are suspect and must be recalled/quarantined.
Join 1,000+ users passing the Sterile Processing Technician
Q11.Which Chemical Indicator is used to differentiate between processed and unprocessed items (external indicator)?
A.Class 1B.Class 3C.Class 5D.Class 6A. Class 1Explanation: Class 1 Process Indicators (like autoclave tape) change color to show the item has been exposed to the sterilization process, preventing unprocessed items from being used.
Join 1,000+ users passing the Sterile Processing Technician
Q12.What information does a 'Load Control Label' provide?
A.The sterilization method, date, sterilizer number, and load numberB.The patient's name and room numberC.The cost of the instrument setD.The manufacturer's phone numberA. The sterilization method, date, sterilizer number, and load numberExplanation: Load labels (Lot numbers) provide the traceability needed to recall specific items if a sterilization failure occurs.
Join 1,000+ users passing the Sterile Processing Technician
Q13.When incubating a Biological Indicator, what is the purpose of the 'Control' vial?
A.To calibrate the incubatorB.To prove the spores were alive/viable before the cycle and the incubator is functioning correctlyC.To compare color changeD.To test the sterilizerB. To prove the spores were alive/viable before the cycle and the incubator is functioning correctlyExplanation: The control vial is not sterilized. It MUST grow bacteria (turn yellow/positive). If the control doesn't grow, the test is invalid because the spores might have been dead to begin with.
Join 1,000+ users passing the Sterile Processing Technician
Q14.Implantable devices should not be released for use until:
A.The cycle is finishedB.The physical monitor printout is checkedC.The Biological Indicator result is known (negative)D.The surgeon calls for itC. The Biological Indicator result is known (negative)Explanation: Because implants remain in the body, the risk of infection is higher. Loads containing implants should be quarantined until the BI confirms sterilization success.
Join 1,000+ users passing the Sterile Processing Technician
Q15.A Class 6 'Emulating Indicator' verifies:
A.That the package is sterileB.That all critical parameters for a *specific* cycle (e.g., 270°F for 4 min) were metC.External exposure onlyD.Air removal onlyB. That all critical parameters for a *specific* cycle (e.g., 270°F for 4 min) were metExplanation: Class 6 indicators are cycle-specific. They emulate the performance of the specific cycle they are designed for (temperature/time) with a high degree of precision.
Join 1,000+ users passing the Sterile Processing Technician
Q16.Testing the sharpness of scissors or the alignment of hemostats is a form of:
A.SterilizationB.DecontaminationC.Quality Assurance / InspectionD.PackagingC. Quality Assurance / InspectionExplanation: Inspection and functional testing ensure the device will perform as intended. Sending up a broken instrument is a quality failure.
Join 1,000+ users passing the Sterile Processing Technician
Q17.Which organization develops the nationally recognized standards (like ST79) for sterilization?
A.OSHAB.FDAC.AAMI (Association for the Advancement of Medical Instrumentation)D.EPAC. AAMI (Association for the Advancement of Medical Instrumentation)Explanation: AAMI sets the consensus standards for sterilization processing. OSHA regulates safety; FDA regulates device manufacturing/sales.
Join 1,000+ users passing the Sterile Processing Technician
Q18.Temperature and humidity logs in the sterile storage area should be checked and recorded:
A.HourlyB.DailyC.WeeklyD.MonthlyB. DailyExplanation: Daily monitoring ensures environmental conditions remain within safe limits to maintain sterility.
Join 1,000+ users passing the Sterile Processing Technician
Q19.During a recall, the first step is to:
A.Call the patientsB.Identify all items processed in the failed load using lot control numbersC.Throw away the sterilizerD.Re-sterilize everything in the departmentB. Identify all items processed in the failed load using lot control numbersExplanation: Traceability is key. Use the lot numbers to find exactly which items were in the failed load and retrieve them.
Join 1,000+ users passing the Sterile Processing Technician
Q20.Which test is used to verify the efficacy of the mechanical washer's cleaning ability (cavitation or impingement)?
A.Bowie-Dick TestB.TOSI (Test Object Surgical Instrument) or foil testC.Geobacillus stearothermophilusD.Class 5 IntegratorB. TOSI (Test Object Surgical Instrument) or foil testExplanation: Cleaning verification tests like TOSI (dried blood soil simulation) or aluminum foil tests verify that the washing equipment is actually removing soil.
Join 1,000+ users passing the Sterile Processing Technician
Q21.Sterilization records (printouts, BI results) must be retained for how long?
A.1 yearB.As specified by local/state statutes or facility policy (often 3+ years)C.Until the patient is dischargedD.ForeverB. As specified by local/state statutes or facility policy (often 3+ years)Explanation: Retention requirements vary by state, but records are typically kept for several years to align with statutes of limitations for lawsuits.
Join 1,000+ users passing the Sterile Processing Technician
Q22.If the physical monitor (printer) on the sterilizer indicates a failure (e.g., temperature dropped), but the chemical indicator passed, what should be done?
A.Release the load based on the CIB.Consider the load unsterile and reprocessC.Ask the surgeonD.Run a BI testB. Consider the load unsterile and reprocessExplanation: The physical monitor is the first and most critical indicator. If the printout shows parameters were not met, the load is unsterile, regardless of what the CIs say.
Join 1,000+ users passing the Sterile Processing Technician
Q23.Using a rigid container that has not been validated for use in your specific sterilizer cycle is an example of:
A.Off-label useB.ValidationC.VerificationD.Quality controlA. Off-label useExplanation: Using a device in a manner not approved by the manufacturer (IFU) is 'off-label' and creates liability/safety risks.
Join 1,000+ users passing the Sterile Processing Technician
Q24.Which of the following is considered a 'Sentinel Event' in sterilization?
A.A torn wrapper found in storageB.A biological indicator turning yellow (positive)C.Sending a non-sterile instrument to the OR that is used on a patientD.The sterilizer running out of paperC. Sending a non-sterile instrument to the OR that is used on a patientExplanation: A sentinel event is an unexpected occurrence involving death or serious physical injury. Using non-sterile instruments on a patient fits this criteria (Joint Commission).
Join 1,000+ users passing the Sterile Processing Technician
Q25.The recommended incubation temperature for Geobacillus stearothermophilus is:
A.35°C - 39°C (95°F - 102°F)B.55°C - 60°C (131°F - 140°F)C.98.6°F (37°C)D.Room temperatureB. 55°C - 60°C (131°F - 140°F)Explanation: Geobacillus stearothermophilus is a thermophile (heat-loving). It requires high incubation temperatures (55-60°C) to grow.
Join 1,000+ users passing the Sterile Processing Technician
Q26.A chemical indicator placed inside a rigid container is called an:
A.External indicatorB.Internal indicatorC.Bowie-DickD.Equipment monitorB. Internal indicatorExplanation: Internal indicators verify that the sterilant penetrated the barrier (container) and reached the instruments inside.
Join 1,000+ users passing the Sterile Processing Technician
Q27.If wet packs are frequently found in a specific sterilizer, the first step in troubleshooting should be:
A.Call the repair technician immediatelyB.Review loading procedures (metal mass, spacing, tilting)C.Increase drying time to 2 hoursD.Use different wrappersB. Review loading procedures (metal mass, spacing, tilting)Explanation: Most wet packs are caused by operator error (improper loading). Verify procedures before assuming equipment failure.
Join 1,000+ users passing the Sterile Processing Technician
Q28.Which agency requires Safety Data Sheets (SDS) be available for all chemicals used in the department?
A.CDCB.OSHAC.FDAD.AAMIB. OSHAExplanation: OSHA's Hazard Communication Standard ('Right to Know') mandates access to SDS for employee safety.
Join 1,000+ users passing the Sterile Processing Technician
Q29.Product quality assurance testing of a rigid sterilization container includes checking the:
A.Color of the lidB.Integrity of the gasket and retention plate/filter mechanismsC.Weight of the empty containerD.Brand nameB. Integrity of the gasket and retention plate/filter mechanismsExplanation: Routine inspection must verify the gasket seals tight and the filter retention mechanism is not bent or loose, ensuring a sterile barrier.
Join 1,000+ users passing the Sterile Processing Technician
Q30.For quality assurance, a Process Challenge Device (PCD) for EtO sterilization should contain:
A.A Bowie-Dick testB.A biological indicator (Bacillus atrophaeus) and a chemical indicatorC.A thermometerD.A syringeB. A biological indicator (Bacillus atrophaeus) and a chemical indicatorExplanation: EtO PCDs typically contain a BI (B. atrophaeus), a CI, and materials (like a syringe or tubing) that create a resistance challenge similar to the load.
Join 1,000+ users passing the Sterile Processing Technician
Q31.Failure to remove soil from the box lock of a hemostat during cleaning is a failure of:
A.SterilizationB.DecontaminationC.PackagingD.TransportB. DecontaminationExplanation: Cleaning is the core of decontamination. If soil remains, sterility cannot be achieved.
Join 1,000+ users passing the Sterile Processing Technician
Q32.A rapid-readout Biological Indicator uses what technology to detect spores?
A.Visual color change (pH shift) onlyB.Fluorescence (enzyme detection)C.SmellD.MicroscopeB. Fluorescence (enzyme detection)Explanation: Rapid-readout BIs (1-3 hours) detect fluorescence produced by enzymes released by living spores, providing results much faster than visual pH color change (24-48 hours).
Join 1,000+ users passing the Sterile Processing Technician
Q33.What is the correct placement for a Bowie-Dick test pack?
A.In a full load, bottom shelfB.In an empty chamber, bottom shelf, over the drainC.In an empty chamber, top shelfD.In a full load, centerB. In an empty chamber, bottom shelf, over the drainExplanation: The test pack must be placed in the most challenging area (over the drain) in an empty chamber to accurately test the vacuum system's ability to remove air.
Join 1,000+ users passing the Sterile Processing Technician
Q34.Verification of the ultrasonic cleaner's cavitation energy should be performed:
A.AnnuallyB.At least weekly (or preferably daily)C.NeverD.When it makes a loud noiseB. At least weekly (or preferably daily)Explanation: Routine verification (weekly/daily) ensures the transducers are functioning and cleaning is effective.
Join 1,000+ users passing the Sterile Processing Technician
Q35.Which document is the ultimate authority on how to process a specific medical device?
A.AAMI ST79B.The textbookC.The Device Manufacturer's Instructions for Use (IFU)D.The Manager's opinionC. The Device Manufacturer's Instructions for Use (IFU)Explanation: The Manufacturer's IFU is the primary source of truth. Validated instructions must be followed to ensure safety and efficacy.
Join 1,000+ users passing the Sterile Processing Technician
Q36.A load record indicates that a 'Control' Biological Indicator was NOT run with the daily BI test. The test BI came back negative (growth). What is the status of the sterilizer?
A.The sterilizer is functioning correctlyB.The test is invalid and inconclusiveC.The sterilizer failedD.The load must be recalledB. The test is invalid and inconclusiveExplanation: Without a positive Control BI, you cannot prove the spores were alive to begin with or that the incubator is working. A negative test result is meaningless (invalid) without a positive control for comparison.
Join 1,000+ users passing the Sterile Processing Technician
Q37.Which scenario represents a 'sentinel event'?
A.A sterilizer breaks downB.A patient develops a deep surgical site infection resulting in loss of limb due to non-sterile instrumentsC.A technician is lateD.A wrapper is tornB. A patient develops a deep surgical site infection resulting in loss of limb due to non-sterile instrumentsExplanation: A sentinel event is an unexpected occurrence involving death or serious physical/psychological injury. A severe infection from dirty instruments qualifies.
Join 1,000+ users passing the Sterile Processing Technician
Q38.When testing a rigid sterilization container, a technician places a biological indicator inside but forgets the chemical indicator. The BI is negative (pass). Can the tray be used?
A.Yes, the BI passedB.No, a chemical indicator is required inside every package to verify exposureC.Yes, if the external lock is intactD.No, the container must be thrown awayB. No, a chemical indicator is required inside every package to verify exposureExplanation: Every package must contain an internal chemical indicator (Class 3, 4, 5, or 6) to verify that sterilant penetrated the package. Without it, the package is considered non-compliant/unsterile.
Join 1,000+ users passing the Sterile Processing Technician